Ti100, Ti110, Ti125, TiR110, TiR125 Thermal Imagers Users Manual PN 3889825 September 2011 © 2011 Fluke Corporation. All rights reserved. Printed in USA. Specifications are subject to change without notice. All product names are trademarks of their respective companies.
LIMITED WARRANTY AND LIMITATION OF LIABILITY Each Fluke product is warranted to be free from defects in material and workmanship under normal use and service. The warranty period is two years and begins on the date of shipment. Parts, product repairs, and services are warranted for 90 days.
Table of Contents Title Page Introduction ............................................................................... How to Contact Fluke ................................................................ Safety Information ..................................................................... Accessories ................................................................................ Before You Start ........................................................................ How to Charge the Battery ...
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual Emissivity Adjustment ........................................................ Adjust by Number........................................................... Select by Table ............................................................... Reflected Background Temperature Compensation............ Transmission/Transmittance Adjustment ............................ Spot Temperatures .............................................................. Spot Markers ........
List of Tables Table 1. 2. 3. 4. 5. 6. 7. Title Page Symbols ..................................................................................... Accessories ................................................................................ Packing Lists.............................................................................. Features and Controls ................................................................ Overview of Controls ................................................................
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual iv
Introduction The Fluke Ti100, Ti110, Ti125, TiR110, and TiR125 Thermal Imagers (the Product or Imager) are handheld, infrared imaging cameras for use in many applications. These applications include equipment troubleshooting, preventive and predictive maintenance, and building diagnostics. The Ti100 is a generalpurpose Imager. The Ti110 and Ti125 are for industrial-commercial maintenance applications. The TiR110 and TiR125 are optimized for building inspection and diagnostics applications.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual How to Contact Fluke To contact Fluke, call one of the following telephone numbers: • • • • • • USA: 1-800-760-4523 Canada: 1-800-36-FLUKE (1-800-363-5853) Europe: +31 402-675-200 Japan: +81-3-6714-3114 Singapore: +65-738-5655 Anywhere in the world: +1-425-446-5500 Or, visit Fluke's website at www.fluke.com. To register your Product, visit http://register.fluke.com. To view, print, or download the latest manual supplement, visit http://us.fluke.
Thermal Imagers Accessories Table 1 is a list of symbols used on the Imager and in this manual. Table 1. Symbols Symbol Description Symbol U Battery status. Battery charging when animated. i Audio indicator p r Description Connected to ac power. Battery removed. Audio recording associated with the displayed image. Pause recording indicator j IR-PhotoNotes™ indicator Video recording in process k Video file indicator On/Off Symbol. Sleep mode. W Important information.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual Before You Start Carefully unpack the items in Table 3. TiR125 TiR110 Ti125 Ti110 Item Ti100 Table 3.
Thermal Imagers Before You Start How to Charge the Battery Before you use the Imager for the first time, charge the battery for a minimum of two and one-half hours. The battery status shows on the five-segment charge indicator. Note New batteries are not fully charged. Two to ten charge/discharge cycles are necessary before the battery charges to its maximum capacity. To charge the battery, use one of the options that follow: Two-Bay Battery Charger Base 1.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual Keep the Imager attached to the charger until the battery condition icon shows a full charge. If you remove the Imager from the charger before a full charge shows, it may have a reduced run-time. Note When the battery is connected to ac power, or the unit is in video mode, the Sleep Mode/Auto Off feature is disabled automatically. Optional 12 V Vehicle Charger 1. Connect the 12 V adapter into the 12 V accessory socket of the vehicle. 2.
Thermal Imagers Features and Controls Features and Controls Table 4 shows the Imager features and controls. Table 4. Features and Controls 2 3 11 4 8 1 12 9 13 19 14 5 18 10 6 15 16 17 21 20 7 Y TTER T BA 3 AR GER SM AR CH C3 Ti SB R WE PO 22 gju01.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual Table 4. Features and Controls (cont.
Thermal Imagers Primary and Secondary Triggers correctly focused infrared images, the IR-Fusion must always be in proper alignment. 2 ft gju04.eps Figure 1. IR-OptiFlex Focus To operate with IR-OptiFlex focus in manual mode or to fine tune the focus, turn the one-touch focus control in a clockwise or counterclockwise direction. As you turn the focus control, you will see a live thermal image on the display as it changes. When your target comes into focus, it shows a sharper image.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual How to Use the Control Buttons Three function buttons (, , ) and four arrow buttons (, , , and ) are the primary controls. These buttons move the cursor through the menu structure to set the features. Table 5 is an overview of the buttons and their actions. In live Manual Mode, the arrow buttons are always active to adjust Level and Span. Table 5.
Thermal Imagers Image Capture Image Capture Point the imager at the object or area of interest. Make sure that the object is in focus. Pull and release the primary trigger. This will capture and freeze the image. To cancel the captured image, pull the primary trigger again or to return to the Live view. Depending on the selected file format settings, the Imager shows the captured image and a menu bar.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual 3. Push to record up to 60 seconds of audio. The display updates to show the recorded time. 4. Push to pause the recorder. 5. Push when done. 6. Push to review the audio file or to save the audio with the image. Listen to a Voice Record The voice (audio) record replays through the speaker. To playback an .is2 file on the SD memory card: 1. Do the steps in the Review Data Files section on page 33 to see the image on the display. 2. Push . 3.
Thermal Imagers SD Memory Card 2. Pull the trigger to capture the image. The image is now in the buffer and you can save or edit. 3. Push to save the image as a file and go back to the live view. SD Memory Card To eject an SD memory card, push in on the exposed edge of the card and then release. The card should pop partially out after you release it. Carefully pull the card out of the slot. To use the SD memory card, make sure that the write-protect lock is open. See Figure 2.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual Temperature Measurement All objects radiate infrared energy. The quantity of energy radiated is based on the actual surface temperature and the surface emissivity of the object. The Imager senses the infrared energy from the surface of the object and uses this data to calculate an estimated temperature value.
Thermal Imagers Menus Menus The menus, together with the three function buttons (, , ) and arrow buttons, are access for thermal image display, camera features, memory setup, and settings for date, time, language, units, file format, and Imager information. Measurement Menu The Measurement Menu has settings for the calculation and display of radiometric temperature measurement data related to the thermal images.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual Fast Auto/Manual Range Toggle When NOT in a menu mode, push for ½ second to toggle between Auto Range and Manual Range. Fast Auto Rescale When in Manual Range and NOT in a menu mode, push for ½ second to automatically rescale the level and span range for objects in the thermal field of view. This feature operates the Imager in a semi-automatic mode if manual fine re-adjustment of level and span with the arrow buttons is not necessary.
Thermal Imagers Menus Total Imager Range Level Span gju02.eps Figure 3. Level and Span Settings Temperature Span for Manual Operation Mode When in manual mode, the span setting contracts or expands in a selected palette in a temperature range within the total range. See Figure 3. In the live manual mode, the arrow buttons are always available to adjust the level and span. To adjust the temperature span: 1. Push to increase or widen the temperature span. 2.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual If you set a value that is <0.60, W shows on the Imager display with this caution: W Caution: Emissivity <0.6 (see Manual) Note Surfaces with an emissivity of <0.60 make reliable and consistent determination of actual temperatures problematic. The lower the emissivity, the more potential error is associated with the Imager's temperature measurement calculations.
Thermal Imagers Menus Select by Table To select from a list of common materials: 1. 2. 3. 4. 5. 6. 7. 8. 9. Push . Push / to highlight Measurement. Push or to view the menu. Push / to highlight Emissivity. Push or to view the menu. Push / to highlight Select Table. Push or to view the emissivity table. Push / to change the value. Push: • to set the change and go back to the live view. • or to set the change and go back to the previous menu.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual 7. Push: • • • to set the change and go back to the live view. or to set the change and go back to the previous menu. to cancel the change and go back to the live view. Note If the Display is set to Display All, you see the information about current reflected background temperature as BG = xx.x.
Thermal Imagers Menus Spot Temperatures The Spot Temperatures are floating HI and LO temperature indicators that move on the display as the temperature measurements of the image fluctuate. To turn on/off the hot and cold spot indicators: 1. Push . 2. Push / to highlight Measurement. 3. Push or to view the menu. 4. Push / to highlight Spot Temp. 5. Push or to view the menu. 6. Push / to toggle this function ON or OFF. 7. Push: • to set the change and go back to the live view.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual Spot Markers Up to three adjustable, fixed-temperature spot markers are available on the display. You can use these markers to highlight a region before you save the image. The marker selection is set as All Off, One Marker, Two Markers, or Three Markers. To set a Marker: 1. Push . 2. Push / to highlight Measurement. 3. Push or to view the menu. 4. Push / to highlight Markers. 5. Push or to view the menu. 6.
Thermal Imagers Menus Center Box The Center Box feature is an adjustable temperature measurement zone (box) that you can center on the infrared image. This zone (box) expands and contracts to different levels in the infrared image. The zone lets the user see an approximate maximum (MAX), average (AVG), and minimum (MIN) temperature measurement in that area. Note When the Center Box feature is on, the HI/LO Spot Temperature markers do not function.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual Palette The Palette menu lets you change the false-color presentation of the infrared images on display or captured. A variety of palettes are available, depending on the model. Some palettes are more suitable for specific applications and can be set as required. Two different palette presentation modes are available on most models (not applicable for the Ti100).
Thermal Imagers Menus To set a palette: 1. Push . 2. Push / to highlight Measurement. 3. Push or to view the menu. 4. Push / to highlight Image. 5. Push or to view the menu. 6. Push / to highlight Palette. 7. Push or to view the menu. 8. Push / to highlight Standard or Ultra Contrast. 9. Push / to select a palette. 10. Push: • • • to set the change and go back to the live view. or to set the change and go back to the previous menu.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual To set the IR-Fusion mode: 1. Push . 2. Push / to highlight Measurement. 3. Push or to view the menu. 4. Push / to highlight Image. 5. Push or to view the menu. 6. Push / to highlight IR-Fusion. 7. Push or to view the menu. 8. Push / to highlight an option. 9. Push: • to set the change and go back to the live view. • or to set the change and go back to the previous menu.
Thermal Imagers Menus To view the Color Alarm menu: 1. 2. 3. 4. 5. Push . Push / to highlight Image. Push or to view the menu. Push / to highlight Color Alarm. Push or to view the menu. Set High-Temperature Color Alarm To set a high-temperature color alarm: 1. 2. 3. 4. From the Color Alarm menu, push / to highlight the option: Set High Alarm. Push to open the Color Alarm menu. Push / to adjust the temperature setting. Push: • to set the change and go back to the live view.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual To set an outside/inside isotherm color alarm: 1. 2. From the Color Alarm menu, push / to highlight Outside or Inside. Push: • • • to set the change and go back to the live view. or to set the change and go back to the previous menu. to cancel the change and go back to the live view. Display Graphics Presentation The options for how you view the on-screen graphics are in the Display menu.
Thermal Imagers Menus Compass The Imager (not applicable for Ti100) includes an eight-point cardinal compass on the display. The compass has on and off functions. This compass lets you accurately record the target locations for analysis and reports. To set the compass: 1. Push . 2. Push / to highlight Camera. 3. Push or to view the menu. 4. Push / to highlight Compass. 5. Push or to view the menu. 6. Push / to highlight ON or OFF. 7. Push to set the option. 8.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual 2. 3. 4. 5. 6. 7. 8. Push / to highlight Camera. Push or to view the menu. Push / to highlight Video. Push or to view the menu. Push / to highlight an option. Push to set the option. Push: • to set the change and go back to the live view. • or to set the change and go back to the previous menu. • to cancel the change and go back to the live view. Video Recording To record: 1. Pull the primary trigger to start recording.
Thermal Imagers Menus 7. During playback, push or for fast forward and rewind. Push to continue normal playback. 8. Push to exit the playback mode. Laser Pointer The laser pointer is a sighting aid and is offset from the infrared camera. As a result, it may not always represent the exact center of the infrared or visible image. The laser dot does not appear on an infrared-only image, but does on visibleonly or AutoBlend images.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual Torch The torch illuminates darker work areas. When set, pull the secondary trigger to operate. Note When the torch is on and an image is captured, the torch momentarily shines brighter and functions as a visible camera flash. To set: 1. Push . 2. Push / to highlight Camera. 3. Push or to view the menu. 4. Push / to highlight Laser/Torch. 5. Push or to view the menu. 6. Push / to highlight an option. 7. Push to set the option. 8.
Thermal Imagers Menus 7. Push: • to set the change and go back to the live view. • or to set the change and go back to the previous menu. • to cancel the change and go back to the live view. Memory Menu The Memory Menu allows the user to review captured images and videos, as well as audio annotations and IR-PhotoNotes in a thumbnail view format. Review Data Files To view stored images on the SD memory card: 1. Push . 2. Push / to highlight Memory. 3.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual 3. Push . 4. Highlight All Image and push . The Imager prompts you to continue or cancel. 5. Push to delete all files on the SD memory card. Settings Menu The Settings menu has adjustments for user preferences such as units of temperature measurement, file format of stored data, auto off settings, date, time, and language.
Thermal Imagers Menus 3. Push or to view the menu. 4. Push / to highlight File Format. 5. Push or to view the menu. 6. Push / to highlight an option. 7. Push to set the option. 8. Push: • to set the change and go back to the live view. • or to set the change and go back to the previous menu. • to cancel the change and go back to the live view. Images saved in the .
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual To set or disable the Auto Off feature: 1. Push . 2. Push / to highlight Settings. 3. Push or to view the menu. 4. Push / to highlight Auto Off. 5. Push or to view the menu. 6. Push / to highlight an option. 7. Push to set the option. 8. Push: • to go back to the live view. • or to set the change and go back to the previous menu. • to cancel the change and go back to the live view.
Thermal Imagers Menus 13. Push: • to set the change and go back to the live view. • or to set the change and go back to the previous menu. • to cancel the change and go back to the live view. Time To set the time: 1. 2. 3. 4. 5. Push . Push / to highlight Settings. Push or to view the menu. Push / to highlight Time. Push or to view the menu. Time displays in two different formats: 24 hour or 12 hour. To set the time format: 1. 2. 3. 4. 5. Push / to highlight time format.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual 4. Push / to highlight Language. 5. Push or to view the menu. 6. Push or to change the setting. 7. Push to set a new language. 8. Push: • to set the change and go back to the live view. • or to set the change and go back to the previous menu. • to cancel the change and go back to the live view. Imager Information You can access information about the Imager from the Settings Menu.
Thermal Imagers Maintenance Maintenance The Imager does not require maintenance. W* Warning To prevent eye damage and personal injury, do not open the Product. The laser beam is dangerous to eyes. Have the Product repaired only through an approved technical site. How to Clean the Case Clean the case with a damp cloth and a weak soap solution. Do not use abrasives, isopropyl alcohol, or solvents to clean the case or lens/window.
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual To get the best performance from the lithium-ion battery, use the guidelines that follow: • Do not store the Imager on the charger for more than 24 hours as reduced battery life may result. • Charge the Imager for a two-hour minimum at six-month intervals for maximum battery life. Without use, the battery will self-discharge in approximately six months. Batteries stored for long periods will need two to ten charging cycles for full capacity.
Thermal Imagers General Specifications User-definable Markers (SmartView only for Ti100) User-selectable Spot Temperatures on the image User-selectable Sleep/Auto Off Color Alarms High Temperature ........................................... Ti110, Ti125, TiR125 Low Temperature (Dewpoint) ......................... Ti125, TiR110, TiR125 Isotherm .......................................................... Ti125, TiR125 User-selectable Center Box (min/avg/max) ........ Ti125, TiR125 ® Software...................
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual Warranty .................................................................2 years Calibration Cycle ...................................................2 years (assumes normal operation and normal aging) Supported Languages...........................................
Thermal Imagers Detailed Specifications Image Presentation Standard Palettes Ti100 ............................................................... Ironbow, Blue-Red, Grayscale, Amber Ti110, Ti125, TiR110, TiR125 ........................ Blue-Red, Grayscale, Inverted Grayscale, High Contrast, Amber, Inverted Amber, Hot Metal, Ironbow Ultra Contrast™ Palettes Ti110, TiR110 ................................................. Ironbow, Blue-Red, Grayscale Ti125, TiR125 .............................................
Ti100, Ti110, Ti125, TiR110, TiR125 Users Manual Image and Data Storage Image Capture, Review, Save Mechanism .........One-handed image capture, review, and save capability Storage Medium ..................................................SD Memory Card (included 2 GB memory card will store at least 1200 fully radiometric (.is2) IR and linked visual images each with 60 seconds voice annotations or 3000 basic (.bmp or .